Salt and pepper artifact in QSM #41
-
Hi, I am trying to use sepia for our study and am running into an issue where there is salt and pepper artifact in the final QSM map. What leads to the artifact appears somewhere at or before phase unwrapping (e.g echo combining, phase unwrapping, or possibly during the data collection steps) and occurs with multiple toolboxes and on multiple subjects we have scanned. It appears at the boundary where phase wrapping occurs. We use 3T MRI and a multi-echo acquisition. I wanted to reach out to see if a better QSM could be generated with our data and to try and find out what may be causing the artifact. I have attached a few screenshots of the final output and intermediate outputs. part-phase_unwrapped:field_maplocal_fieldQSMPlease let me know if I can provide any additional information. |
Beta Was this translation helpful? Give feedback.
Replies: 3 comments 5 replies
-
Hi James, It seems also to me that the artefact appears in the total field estimation process. The artefact seems to be caused by low SNR in certain brain regions after echo combination/total field estimation and subsequently propagates to the downstream processing steps. To have a better insight of the artefact origin, it would be helpful if you can also post the images of the original multi-echo wrapped phase and magnitude data (e.g. same slices for each echo) here, so that I can have a look into data quality. Can you also let me know which echo combination method and phase unwrapping method used? |
Beta Was this translation helpful? Give feedback.
-
Thanks for sharing the images! The low SNR issue is caused by the so-called "phase singularity (or open-ended fringe line) in the phase data (as indicated by the arrows). This is the result of sub-optimal coil combination on the phase data, where the differences in phase offset of each coil image cancel each other (destructive interference) in the reconstructed data when the multi-coil data are combined. If by any chance, the raw k-space data or multi-channel images are still available then there might be a way to correct it with the acquired data. However, I'm not aware of any software packages available to correct this issue with the coil combined data (since the offsets are coil-dependent and this information is lost in the reconstructed data). Would you mind also letting me know which scanner and software version the data were acquired? I think most of the vendors support combining phase information from the RF array coils appropriately. I might be able to provide more information about how the phase images can be exported correctly for future data acquisition. |
Beta Was this translation helpful? Give feedback.
-
Just an additional point for future acquisitions, in the VE version and if
using the product gradient echo sequence, you can perform acceleration in
only one dimension. If using the vibe sequence you accelerate in multiple
dimensions and will not get these artefacts.
…On Thu, Sep 22, 2022 at 10:25 AM Kwok-Shing Chan ***@***.***> wrote:
In this case I'm afraid there is not much we could do on the acquired
data. I hope that the information could at least be useful for the future
data acquisition.
—
Reply to this email directly, view it on GitHub
<#41 (reply in thread)>,
or unsubscribe
<https://github.com/notifications/unsubscribe-auth/ACELXMGRW7KQSGLWHFFW7ATV7QJXDANCNFSM6AAAAAAQNWSXV4>
.
You are receiving this because you are subscribed to this thread.Message
ID: ***@***.***>
|
Beta Was this translation helpful? Give feedback.
Thanks for sharing the images!
The low SNR issue is caused by the so-called "phase singularity (or open-ended fringe line) in the phase data (as indicated by the arrows). This is the result of sub-optimal coil combination on the phase data, where the differences in phase offset of each coil image cancel each other (destructive interference) in the reconstructed data when the multi-coil data are combined. If by any chance, the raw k-space data or multi-channel images are still available then there might be a way to correct it with the acquired data. However, I'm not aware of any software packages available to correct this issue with the coil combined data (since the offsets are coil-depend…